AlVa_264671
New Contributor II

Hi

I'd like to use the ADC_SAR  and I have some question regarding using this API  :  void ADC_SetScaledGain(int16 adcGain)

According to the data sheet     adcGain = counts x 10 / Vmeasured.

This work OK , I have a prototype working based on that..

Now, assume I want to calibrate the signal chain including gain op amp and active filter and also assume that I cannot use the Delta

sigma ADC.

I can supply a known voltage  close to the reference voltage (in my case 1024 mv.)  and measure as directed

with a voltmeter  the voltage at the ADC input an get the value of adcGain.

If I wanted to make a PCB serially I would need to do this measurement on each board. ?

How automate this? Some suggestion ?

Thanks in advance

0 Likes
1 Solution
odissey1
Honored Contributor II

AIVa,

Instead of factory calibration against an external voltage (e.g. 1024mV) and adjusting Gain setting, consider using onboard reference Vref = 1.02V as input test voltage, and record measured value (e.g. Vmeasured = 0.99V). Then all data need to be corrected for coefficient: coeff = Vref/Vmeasured. This procedure can be run every time on startup (with some MUX), so no factory calibration required.

High-precision absolute voltage measurement can be tricky if better than 10^-3 resolution is needed. Usually, a stable external voltage source is used for calibration (like Mercury-Silver battery) on every measurement, as internal Vref drifts away. If accuracy of few mV is sufficient, then existing PSoC factory calibration of ADC should suffice.

/odissey1

View solution in original post

0 Likes
3 Replies
odissey1
Honored Contributor II

AIVa,

Instead of factory calibration against an external voltage (e.g. 1024mV) and adjusting Gain setting, consider using onboard reference Vref = 1.02V as input test voltage, and record measured value (e.g. Vmeasured = 0.99V). Then all data need to be corrected for coefficient: coeff = Vref/Vmeasured. This procedure can be run every time on startup (with some MUX), so no factory calibration required.

High-precision absolute voltage measurement can be tricky if better than 10^-3 resolution is needed. Usually, a stable external voltage source is used for calibration (like Mercury-Silver battery) on every measurement, as internal Vref drifts away. If accuracy of few mV is sufficient, then existing PSoC factory calibration of ADC should suffice.

/odissey1

View solution in original post

0 Likes
AlVa_264671
New Contributor II

Odissey1

Thank you.

You said

record measured value (e.g. Vmeasured = 0.99V).

Can you clarify how you get those (in this case )  0.99V ?

1- Using the same SAR_ADC ?

2- Reading the counts  you get from 1.02 V ?

0 Likes
odissey1
Honored Contributor II

AIVa,

The values Vmeasired = "0.99V" and Vref="1.02V" are purely artificial numbers out of my head. They were provided just to clarify the idea.

On startup, lets switch mux to a pin, connected to the PSoC reference Vref=1.02V and measure signal using ADC_SAR (lets call it "Vref"). The same Vref should be used as ADC_SAR reference voltage. Ideally, if SAR-ADC is calibrated perfectly, the value returned should be 1.024V, but in reality it can be slightly off (lets say it is Vmeasured = 0.99V). This discrepancy may come from uncalibrated SAR_ADC. Now we know that every measurement returned by the SAR_ADC is off by a coefficient  0.99V/1.024V  = 0.96679. Therefore, to get correct ADC values, all ADC_SAR results must be scaled up by 1.024/0.99 = 1.0343434.

As alternative procedure, you may adjust ADC_SAR Gain calibration by iterations, until measured result will become 1.024V, but i think that it is slower.

/odissey1

0 Likes