I'm curious as to whether I could use two iDAC to create a constant current biphasic pulse to stimulate biological tissue. I don't completely understand the iDAC compliance voltage and how it would relate to my project. I would like to produce a 100 microAmp pulse with a total duration of 1 ms, and the load will be 0.5-1.0 MegaOhms.
Cound someone comment on the iDAC compliance voltage, and if my load and current requirements would be possible using the iDAC.
Labels
- Labels:
-
PSoC 3
1 Reply
Anonymous
Not applicable
Jan 31, 2012
06:31 PM
- Mark as New
- Bookmark
- Subscribe
- Mute
- Subscribe to RSS Feed
- Permalink
- Report Inappropriate Content
Jan 31, 2012
06:31 PM
100uA into 1Mohm means 100V across the load. Don't think you can use the pSoC IDAC directly. you need to have external circuitry to do this.