Hi, I am developing a medicine application which includes a programmed Cypress CY8C27443-24PXi integrated circuit. I have been trying to use it on a protoboard instead of the PSoC itself for space purpose. I have found out that, after programming it and putting it on my protobard to test it, it does not work. I have put the power lines properly and checked the pin out diagram for the correct input and output ports but it seems like it does not want to work at all. I put it back to the PSoC and connect it to my protoboard with some jumpers and works just right. Is there something that needs to be considered for the IC to work out of the plataform on a protoboard or on a PCB desing?
Any help would be appreciated.
Thanks.
Juan Luis Gantus Bernés.
9th. Semester Student of Biomedical Engieering.
Solved! Go to Solution.
- Labels:
-
PSoC 1
- Mark as New
- Bookmark
- Subscribe
- Mute
- Subscribe to RSS Feed
- Permalink
- Report Inappropriate Content
There is nothing secret / extra to run a PSoC when it is programmed, so you will have to concentrate on the differences between your two configurations.
Power supply: Voltage, Noise
Bad contacts
Caps
Bob
- Mark as New
- Bookmark
- Subscribe
- Mute
- Subscribe to RSS Feed
- Permalink
- Report Inappropriate Content
There is nothing secret / extra to run a PSoC when it is programmed, so you will have to concentrate on the differences between your two configurations.
Power supply: Voltage, Noise
Bad contacts
Caps
Bob
- Mark as New
- Bookmark
- Subscribe
- Mute
- Subscribe to RSS Feed
- Permalink
- Report Inappropriate Content
I have and still not working. I have double checked the pin out diagram and the input and outputs thatI declare and proceed to connect it to the next stage of my circuit and yet still working wrong. I put pin number 28 (VDD) to 5V and pin number 14(VSS) to ground and use the other declared ports as inputs and outputs but the resulting signal is not what I should get (Already got it on the PSoC board).
- Mark as New
- Bookmark
- Subscribe
- Mute
- Subscribe to RSS Feed
- Permalink
- Report Inappropriate Content
There is no way out, there is an error in your board.
I asked for the noise on your 5V supply. How much is it?
Are you using an x-tal on any of your boards?
Bob
- Mark as New
- Bookmark
- Subscribe
- Mute
- Subscribe to RSS Feed
- Permalink
- Report Inappropriate Content
The 5V power supply is just fine, I am using a medical grade power supply and regarding the x-tal the answer is no, I am not using one, I am using the IC internal oscillator. I will keep testing it but for now it is just not working on my protoboard. Thanks for the help though.
- Mark as New
- Bookmark
- Subscribe
- Mute
- Subscribe to RSS Feed
- Permalink
- Report Inappropriate Content
" I am using a medical grade power supply" This is a quality guarant for man-safe design, but not for low noise.
"Anything that is not expressed with numbers is a pure opinion" (Robert A Heinlein)
Bob
- Mark as New
- Bookmark
- Subscribe
- Mute
- Subscribe to RSS Feed
- Permalink
- Report Inappropriate Content
You have a bypass on the protoboard of typically 10 uF
electrolytic and a .1 uF ceramic ? Next to Vdd lead.
The grounds are commoned to one rail on the board ?
You have looked at Vdd with scope, and see < ~ 200 mV noise pk - pk ?
Scope if DSO put on infinite persistence when you do this.
Post a picture of board and wiring.
Regards, Dana.
- Mark as New
- Bookmark
- Subscribe
- Mute
- Subscribe to RSS Feed
- Permalink
- Report Inappropriate Content
Thanks, Dana. I checked everything. Now my circuit is working as it is supposed to.
BTW, Bob, a medical grade power supply is indeed a man-safe guarantee but it's also accurate with its technical specifications since it is made for medical applications. Thanks for not helping, though!
Problem solved.
- Mark as New
- Bookmark
- Subscribe
- Mute
- Subscribe to RSS Feed
- Permalink
- Report Inappropriate Content
I had occasion as a field engineer to represent several companies
making medical grade power supplies. Noise, conducted and radiated,
was indeed very much a concern and speced. In light of the issues
surrounding EKG, EEG, and other sensor technologies, a lot of attention
and specification covered the topic of noise, shielding, transmission,
etc..
I worked (as FAE) on portable MRI machine where an extensive, beyond datasheet,
charcterization was done and manufacturer advised of changes to upgrade their
already low noise supplies. Capacitor technologies were also extensively examined.
As well as LCD display emitted noise and B/L inveter technology.
Regards, Dana.