Hi, for my college thesis I'm trying to implemet ecg amplifer without any external components(!!!), and I need some help. I'm using PSoC 5lp cy8ckit-059. I found this picture of ecg amplifier circuit somewhere (attached 1st pic). First two PGAs are instrumentation amplifier. I also found that for instrumentation amplifier PGA gain has to be 2. (2nd pic). But I have to achieve amplification somewhere aroud 1000. So if i put the 3rd PGA (after filters) G=50, with first two PGAs (before filter) G=2, where can i squeeze the rest needed amplificaiton=5 (or 4) so the whole amplification is 2*2*50*5=1000 (or 2*2*50*4=800)?
Thank you
Solved! Go to Solution.
Labels
- Labels:
-
ispn:39618:1:0
-
l1:314:1:0
-
PSoC5LP
1 Solution
Mar 23, 2017
03:30 PM
- Mark as New
- Bookmark
- Subscribe
- Mute
- Subscribe to RSS Feed
- Permalink
- Report Inappropriate Content
Mar 23, 2017
03:30 PM
And why ECG needs gain of 1000? Let's calculate resolution at the input of 20-bit DelSig_ADC in differential more with buffer scale set to V(-) +/-.125Vref: V=(+/-) 0.125 * 1.024V / 2^20 = 0.12uV. This is very high precision without using any amplifier in front of ADC, input noise will be much higher than that. Should you use ADC_SAR, OpAmp would be appropriate.