Bridge Control Panel software???

Tip / Sign in to post questions, reply, level up, and achieve exciting badges. Know more

cross mob
David_Zhang
Level 5
Level 5
Distributor - Arrow(GC)
First like received 50 sign-ins 25 sign-ins

  How to use Bridge control panel to communicate with PSOC;

I can't find the instructions text files in CYPESS web;

such as:

    i don't know +,p what it means and why there are three byte date of ff after a4;  please help, thanks;

pastedImage_1.png

0 Likes
1 Solution
ShanmathiN_06
Employee
Employee
100 replies posted 50 replies posted 25 replies posted

Hi David,

You can navigate to help section (F1) of Bridge control panel and select communication.

Various symbols used in commands and the interpretation of results is available.

P -> Generates a stop condition on the bus

+ -> Append two commands together

It looks like you are reading 4 bytes of data from the I2C slave device.

So, ensure your slave is also configured to send out 4 bytes of data.

Thanks,
Shanmathi